Teledyne t320 Carbon Monoxide Alarm User Manual


 
Teledyne API – Model T300/T300M CO Analyzer Overview of Operating Modes
95
4.3. SETUP MODE
The SETUP mode contains a variety of choices that are used to configure the analyzer’s
hardware and software features, perform diagnostic procedures, gather information on
the instrument’s performance, and configure or access data from the internal data
acquisition system (DAS). For a visual representation of the software menu trees, refer
to Appendix A.
Setup Mode is divided between Primary and Secondary Setup menus and can be
protected through password security.
4.3.1. PASSWORD SECURITY
Setup Mode can be protected by password security through the SETUP>PASS menu
(Section 5.5) to prevent unauthorized or inadvertent configuration adjustments.
4.3.2. PRIMARY SETUP MENU
4.3.3. THE AREAS ACCESSIBLE UNDER THE Setup MODE ARE SHOWN IN
TABLE 4-4 AND SECONDARY SETUP MENU (SETUP>MORE)
Table 4-5.
Table 4-4: Primary Setup Mode Features and Functions
MODE OR FEATURE
CONTROL
BUTTON
DESCRIPTION
MANUAL
SECTION
Analyzer Configuration
CFG
Lists button hardware and software configuration information 5
Auto Cal Feature
ACAL
Used to set up and operate the AutoCal feature.
Only appears if the analyzer has one of the internal valve
options installed.
5.2
and
9.4
Internal Data Acquisition
(DAS)
DAS
Used to set up the DAS system and view recorded data 7
Analog Output Reporting
Range Configuration
RNGE
Used to configure the output signals generated by the
instruments Analog outputs.
5.7
Calibration Password Security
PASS
Turns the calibration password feature ON/OFF. 5.3
Internal Clock Configuration
CLK
Used to Set or adjust the instrument’s internal clock. 5.6
Advanced SETUP features
MORE
This button accesses the instruments secondary setup menu.
See
Table 6-5
06864B DCN6314