Teledyne API – Model T300/T300M CO Analyzer Communications Setup and Operation
165
6.7.3.4. SETTING HESSEN PROTOCOL STATUS FLAGS
Teledyne API’s implementation of Hessen protocols includes a set of status bits that the
instrument includes in responses to inform the host computer of its condition. Each bit
can be assigned to one operational and warning message flag. The default settings for
these bit/flags are:
Table 6-6: Default Hessen Status Flag Assignments
STATUS FLAG NAME DEFAULT BIT ASSIGNMENT
WARNING FLAGS
SAMPLE FLOW WARNING 0001
BENCH TEMP WARNING 0002
SOURCE WARNING 0004
BOX TEMP WARNING 0008
WHEEL TEMP WARNING 0010
SAMPLE TEMP WARN 0020
SAMPLE PRESS WARN 0040
INVALID CONC
(The Instrument’s Front Panel Display Will Show The
Concentration As “Warnings”)
0080
OPERATIONAL FLAGS
1
Instrument OFF 0100
In MANUAL Calibration Mode 0200
In ZERO Calibration Mode
4
0400
In O
2
Calibration Mode (if O
2
sensor installed )
2,4
0400
In CO
2
Calibration Mode (if CO
2
sensor installed )
2,4
0400
In SPAN Calibration Mode 0800
UNITS OF MEASURE FLAGS
UGM 0000
MGM 2000
PPB 4000
PPM 6000
SPARE/UNUSED BITS
1000, 8000
UNASSIGNED FLAGS (0000)
AZERO WARN
2
DCPS WARNING
CANNOT DYN SPAN
2
REAR BOARD NOT DET
CANNOT DYN ZERO
3
SYNC WARNING
1
CONC ALARM 1
3
SYSTEM RESET
1
CONC ALARM 2
3
1
These status flags are standard for all instruments and should probably not be modified.
2
Only applicable if the optional internal span gas generator is installed.
3
Only applicable if the analyzer is equipped with an alarm options.
3
It is possible to assign more than one flag to the same Hessen status bit. This allows the grouping of
similar flags, such as all temperature warnings, under the same status bit.
Be careful not to assign conflicting flags to the same bit as each status bit will be triggered if any of the
assigned flags is active.