Calibration Procedures Teledyne API – Model T300/T300M CO Analyzer
220
Offset for CO REF signal
SAMPLE RANGE=50.0 PPM CO= XX.XX
<TST TST> CAL SETUP
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
SETUP X.X ENTER PASSWORD
818 ENTREXIT
Continue pressing NEXT until ...
DIAG OPTIC DARK CALIBRATION
PREV NEXT ENTR EXIT
DIAG DARK CO DARK CALIBRATION
VIEW CAL EXIT
DIAG DARK REF DARK OFFSET: 0.0mV
EXIT
Offset for CO MEAS signal
DIAG DARK MEAS DARK OFFSET: 0.0mV
EXIT
DIAG DARK DARK CAL 1% COMPLETE
EXIT
DIAG DARK DARK CALIBRATION ABORTED
EXIT
Calibration runs automatically
9.6.2. PRESSURE CALIBRATION
A sensor at the sample chamber outlet continuously measures the pressure of the sample
gas. These data are used to compensate the final CO concentration calculation for
changes in atmospheric pressure and is stored in the CPU’s memory as the test function
PRES (also viewable via the front panel). See also Section 5.9.6.
IMPORTANT
IMPACT ON READINGS OR DATA
This calibration must be performed when the pressure of the sample gas
is equal to ambient atmospheric pressure. Before performing the
following pressure calibration procedure, disconnect the sample gas
pump and sample gas-line vent from the rear panel sample gas inlet.
06864B DCN6314