Teledyne t320 Carbon Monoxide Alarm User Manual


 
Getting Started Teledyne API – Model T300/T300M CO Analyzer
46
3.3. CONNECTIONS AND SETUP
This section presents the electrical (Section 3.3.1) and pneumatic (Section 3.3.2)
connections for setup and preparing for instrument operation
3.3.1. ELECTRICAL CONNECTIONS
Note To maintain compliance with EMC standards, it is required that the cable
length be no greater than 3 meters for all I/O connections, which include
Analog In, Analog Out, Status Out, Control In, Ethernet/LAN, USB, RS-232,
and RS-485.
3.3.1.1. CONNECTING POWER
Attach the power cord to the analyzer and plug it into a power outlet capable of carrying
at least 10 A current at your AC voltage and that it is equipped with a functioning earth
ground.
WARNING
ELECTRICAL SHOCK HAZARD
High Voltages are present inside the analyzer’s case.
Power connection must have functioning ground connection.
Do not defeat the ground wire on power plug.
Turn off analyzer power before disconnecting or
connecting electrical subassemblies.
Do not operate with cover off.
CAUTION
GENERAL SAFETY HAZARD
The T300/T300M Analyzer can be configured for both
100-130 V and 210-240 V at either 47 Hz or 63 Hz.
To avoid damage to your analyzer, make sure that the AC power voltage matches
the voltage indicated on the analyzer’s model/specs label (See Figure 3-4) before
plugging the T300/T300M into line power.
06864B DCN6314