12.1.4.3. RELAY BOARD STATUS LEDS
There are eight LEDs located on the Relay Board. The most important of which is D1,
which indicates the health of the I
2
C bus. If D1 is blinking the other faults following
LEDs can be used in conjunction with DIAG menu signal I/O to identify hardware
failures of the relays and switches on the relay (see Section 12.1.3 and Appendix D).
Table 12-4: I
2
C Status LED Failure Indications
LED FUNCTION FAULT STATUS INDICATED FAILURE(S)
D1
(Red)
I
2
C bus Health
(Watch Dog
Circuit)
Continuously ON
or
Continuously OFF
Failed/Halted CPU
Faulty Motherboard, Touchscreen or Relay Board
Faulty Connectors/Wiring between Motherboard,
Touchscreen or Relay Board
Failed/Faulty +5 VDC Power Supply (PS1)
STATUS LED’s
DC VOLTAGE TEST
POINTS
RELAY PCA
PN 04135
Figure 12-5: Relay Board Status LEDs