Teledyne API – Model T300/T300M CO Analyzer Getting Started
77
Table 3-13 lists brief descriptions of the warning messages that may occur during start
up.
Table 3-13: Possible Warning Messages at Start-Up
Message MEANING
ANALOG CAL WARNING
The instrument's A/D circuitry or one of its analog outputs is not calibrated.
BENCH TEMP WARNING
Optical bench temperature is outside the specified limits.
BOX TEMP WARNING
The temperature inside the T300/T300M chassis is outside the specified limits.
CANNOT DYN SPAN
2
Remote span calibration failed while the dynamic span feature was set to
turned on.
CANNOT DYN ZERO
3
Remote zero calibration failed while the dynamic zero feature was set to turned
on.
CONFIG INITIALIZED
Configuration was reset to factory defaults or was erased.
DATA INITIALIZED
DAS data storage was erased.
PHOTO TEMP WARNING
Photometer temperature outside of warning limits specified by
PHOTO_TEMP_SET variable.
REAR BOARD NOT DET
Motherboard was not detected during power up.
RELAY BOARD WARN
CPU is unable to communicate with the relay PCA.
SAMPLE FLOW WARN
The flow rate of the sample gas is outside the specified limits.
SAMPLE PRESS WARN
Sample pressure outside of operational parameters.
SAMPLE TEMP WARN
The temperature of the sample gas is outside the specified limits.
SOURCE WARNING
The IR source may be faulty.
SYSTEM RESET
1
The computer was rebooted.
WHEEL TEMP WARNING
The Gas Filter Correlation Wheel temperature is outside the specified limits.
1
Typically clears 45 minutes after power up.
2
Clears the next time successful zero calibration is performed.
3
Clears the next time successful span calibration is performed.
Table 3-14 lists brief descriptions of the warning messages that may occur during start
up for T300 analyzers with optional second gas options or alarms installed.