Teledyne API – Model T300/T300M CO Analyzer Setup Menu
141
5.10.1. SETTING THE T300 CONCENTRATION ALARM LIMITS
To enable either of the CO concentration alarms and set the limit points, press:
SETUP X.X CO ALRM 2: OFF
ON ENTR EXIT
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG ALRM EXIT
DIAG FCAL CO ALARM 2=300.00 PPM
3 00.00 ENTREXIT
Toggle these
buttons to Set
the alarm point
EXIT discards the new
setting
ENTR accepts the
new setting
SAMPLE RANGE=50.0 PPM CO= XX.XX
<TST TST> CAL SETUP
SETUP X.X CO ALRM 2, DISALBED
NEXT EDIT PRNT EXIT
Toggle this button to
enable/disable
the alarm
Continue pressing NEXT until the desired
Alarm is selected
06864B DCN6314