Setup Menu Teledyne API – Model T300/T300M CO Analyzer
132
An alternative method for measuring the output of the Current Loop converter is to
connect a 250 ohm 1% resistor across the current loop output in lieu of the current
meter (see Figure 3-9 for pin assignments and diagram of the analog output connector).
This allows the use of a voltmeter connected across the resistor to m
easure converter
output as VDC or mVDC.
V
+DC Gnd
Figure 5-6: Alternative Setup Using 250 Resistor for Checking Current Output Signal Levels
In this case, follow the procedure above but adjust the output for the following values:
Table 5-8: Current Loop Output Check
% FS
Voltage across Resistor for
2-20 mA
Voltage across
Resistor for 4-20 mA
0 500 mVDC 1000 mVDC
100 5000 mVDC 5000 mVDC