Teledyne API – Model T300/T300M CO Analyzer Setup Menu
103
5.4.3.2. DUAL RANGE MODE (DUAL)
Selecting the DUAL range mode allows the A1 and A2 outputs to be configured with
different reporting ranges. The analyzer software calls these two ranges low and high.
The LOW range setting corresponds with the analog output labeled A1 on the rear
panel of the instrument.
The HIGH range setting corresponds with the A2 output.
While the software names these two ranges low and high, they do not have to be
configured that way. For example: The low range can be set for a span of 0-1000 ppm
while the high range is set for 0-500 ppm.
In DUAL range mode the RANGE test function displayed on the front panel will be
replaced by two separate functions:
RANGE1: The range setting for the A1 output.
RANGE2: The range setting for the A2 output.
To select the DUAL range mode press following buttonstroke sequence
.
When the instrument’s range mode is set to DUAL, the concentration field in the upper
right hand corner of the display alternates between displaying the low range value and
the high range value. The concentration currently being displayed is identified as
follows: C1= LOW (or A1) and C2 = HIGH (or A2).
IMPORTANT
IMPACT ON READINGS OR DATA
In DUAL range mode the LOW and HIGH ranges have separate slopes
and offsets for computing CO concentrations. The two ranges must be
independently calibrated.
06864B DCN6314