Teledyne API – Model T300/T300M CO Analyzer Theory of Operation
313
Thermo-Electric
Cooler
Control Circuit
Sample &
Hold
Circuits
(x4)
Amplifiers
Dark
Switch
Photo-
detector
Pre Amp
Phase
Lock
Loop
E-Test
Generator
Compact
Programmable
Logic Device
M/R Sensor
Segment
Sensor
E Test Control
Dark Switch
Control
X10 Clock
Phase Lock
E Test A Gate
Dark Test Gate
X1 Reference
Segment
Status LED
Phase Lock Warning
M/R
Status LED
Segment Clock
E Test B Gate
Measure Gate
Measure Dark Gate
Reference Gate
Reference Dark Gate
CO MEAS
CO Reference
56V
Bias
Variable
Gain Amp
From GFC
Wheel
From CPU
via Mother
Board
Signal
Conditioner
Signal
Conditioner
10
x10
TEC Control
PHT DRIVE
Figure 13-13: T300/T300M Sync/Demod Block Diagram
13.4.3.1. SIGNAL SYNCHRONIZATION AND DEMODULATION
The signal emitted by the IR photo-detector goes through several stages of amplification
before it can be accurately demodulated. The first is a pre-amplification stage that raises
the signal to levels readable by the rest of the sync/demod board circuitry. The second is
a variable amplification stage that is adjusted at the factory to compensate for
performance variations of mirrors, detectors, and other components of the optical bench
from instrument to instrument.
The workhorses of the sync/demod board are the four sample-and-hold circuits that
capture various voltage levels found in the amplified detector signal needed to determine
the value of CO MEAS and CO REF. They are activated by logic signals under the
control of a compact Programmable Logic Device (PLD), which in turn responds to the
output of the Segment Sensor and M/R Sensor as shown in Figure 13-9.
06864B DCN6314