Setup Menu Teledyne API – Model T300/T300M CO Analyzer
122
5.9.3.1. ANALOG OUTPUT VOLTAGE / CURRENT RANGE SELECTION
In its standard configuration, each of the analog outputs is set to output a 0–5 VDC
signals. Several other output ranges are available. Each range has is usable from -5% to
+ 5% of the rated span.
Table 5-6: Analog Output Voltage Ranges
RANGE NAME RANGE SPAN MINIMUM OUTPUT MAXIMUM OUTPUT
0.1V
0-100 mVDC -5 mVDC 105 mVDC
1V
0-1 VDC -0.05 VDC 1.05 VDC
5V
0-5 VDC -0.25 VDC 5.25 VDC
10V
0-10 VDC -0.5 VDC 10.5 VDC
The default offset for all VDC ranges is 0-5 VDC.
CURR
0-20 mA 0 mA 20 mA
While these are the physical limits of the current loop modules, typical applications use 2-20 or 4-20 mA for the lower and upper
limits. Please specify desired range when ordering this option.
The default offset for all current ranges is 0 mA.
Current outputs are available only on A1-A3.