Teledyne t320 Carbon Monoxide Alarm User Manual


 
APPENDIX A-2: Setup Variables For Serial I/O Teledyne API - T300/T300M and M300E/EM PN 04906H (DCN5840)
A-18
Setup Variable Numeric
Units
Default
Value
Value
Range
Description
SERIAL_NUMBER —
“00000000 ”
0
Any character
in the allowed
character set.
Up to 100
characters
long.
Unique serial number for
instrument.
DISP_INTENSITY — HIGH
0
HIGH,
MED,
LOW,
DIM
Front panel display intensity.
I2C_RESET_ENABLE — ON OFF, ON
ON enables automatic reset of
the I
2
C bus in the event of
communication failures; OFF
disables automatic reset.
CLOCK_FORMAT —
“TIME=%H:%
M:%S”
Any character
in the allowed
character set.
Up to 100
characters
long.
Time-of-day clock format flags.
Enclose value in double quotes
(") when setting from the RS-232
interface.
“%a” = Abbreviated weekday
name.
“%b” = Abbreviated month name.
“%d” = Day of month as decimal
number (01 – 31).
“%H” = Hour in 24-hour format
(00 – 23).
“%I” = Hour in 12-hour format (01
– 12).
“%j” = Day of year as decimal
number (001 – 366).
“%m” = Month as decimal
number (01 – 12).
“%M” = Minute as decimal
number (00 – 59).
“%p” = A.M./P.M. indicator for
12-hour clock.
“%S” = Second as decimal
number (00 – 59).
“%w” = Weekday as decimal
number (0 – 6; Sunday is 0).
“%y” = Year without century, as
decimal number (00 – 99).
“%Y” = Year with century, as
decimal number.
“%%” = Percent sign.
ALARM_TRIGGER
3,4
Cycles 10 1–100
Concentration alarm trigger
sensitivity adjustment.
REF_SDEV_LIMIT mV 50 0.1–500
Reference detector standard
deviation must be below this limit
to switch out of startup mode.
06864B DCN6314