Calibration Procedures Teledyne API – Model T300/T300M CO Analyzer
216
SETUP X.X STARTING DATE: 04
–
SEP
–
08
<SET SET> EDIT EXIT
SETUP X.X STARTING TIME:00:00
<SET SET> EDIT EXIT
SETUP X.X STARTING TIME:00:00
1 4 : 1 5 ENTR EXIT
Toggle buttons to set
time:
Format : HH:MM
This is a 24 hr clock . PM
hours are 13 – 24.
Example 2:15 PM = 14:15
CONTINUED FROM PREVIOUS PAGE -
STARTING DATE
CONTINUE NEXT PAGE
With DURATION TIME
SETUP X.X STARTING TIME:14:15
<SET SET> EDIT EXIT
SETUP X.X DELTA DAYS: 1
<SET SET> EDIT EXIT
SETUP X.X DELTA DAYS: 1
0 0 2 ENTR EXIT
Toggle buttons to set
number of days between
procedures (1-365).
SETUP X.X DELTA DAYS:2
<SET SET> EDIT EXIT
SETUP X.X DELTA TIME00:00
<SET SET> EDIT EXIT
SETUP X.X DELTA TIME: 00:00
0 0 :3 0 ENTR EXIT
SETUP X.X DELTA TIME:00:30
<SET SET> EDIT EXIT
Toggle buttons to set
delay time for each
iteration of the sequence:
HH:MM
(0 – 24:00)
06864B DCN6314