Teledyne t320 Carbon Monoxide Alarm User Manual


 
Setup Menu Teledyne API – Model T300/T300M CO Analyzer
102
5.4.3.1. SINGLE RANGE MODE (SNGL)
Single Range Mode (SNGL) is the default reporting range mode for the analyzer.
When the single range mode is selected (SNGL), all analog CO concentration outputs
(A1 and A2) are slaved together and set to the same reporting range limits (e.g. 500.0
ppb). The span limit of this reporting range can be set to any value within the physical
range of the analyzer.
Although both outputs share the same concentration reporting range, the electronic
signal ranges of the analog outputs may still be configured for different values (e.g. 0-5
VDC, 0-10 VDC, etc; see Section 5.9.3.1)
To select SNGL range m
ode and to set the upper limit of the range, press:
06864B DCN6314