Teledyne API – Model T300/T300M CO Analyzer Calibration Procedures
227
9.7.2. CO
2
SENSOR CALIBRATION PROCEDURE
Presented here are first the setup and then the calibration steps for the CO
2
Sensor.
9.7.2.1. CO2 PNEUMATICS CONNECTIONS
The pneumati
c connections for calibrating are as follows
Figure 9-8: CO
2
Sensor Calibration Set Up
CO
2
SENSOR ZERO GAS: Teledyne API recommends using pure N
2
when calibration
the zero point of your CO
2
sensor option.
CO
2
SENSOR SPAN GAS: Teledyne API recommends using 16% CO
2
in N
2
when
calibration the span point of your CO
2
sensor option (Table 3-12) is 20%.
06864B DCN6314