Teledyne t320 Carbon Monoxide Alarm User Manual


 
Teledyne API – Model T300/T300M CO Analyzer Getting Started
81
While these are the default setting for the T300/T300M Analyzer, it is recommended
that you verify them before proceeding with the calibration procedure, by pressing:
SETUP X.X RANGE MODE:SINGL
SNGL DUAL AUTO ENTR EXIT
SAMPLE RANGE=50.0 PPM CO= XX.XX
<TST TST> CAL SETUP
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X RANGE CONTROL MENU
MODE SET UNIT DIL EXIT
SETUP X.X RANGE: 50.0 Conc
00050.0ENTREXIT
SETUP X.X CONC UNITS:PPM
PPB PPM UGM MGM ENTR EXIT
SETUP X.X RANGE CONTROL MENU
MODE SET UNIT DIL EXIT
SETUP X.X RANGE CONTROL MENU
MODE SET UNIT DIL EXIT
Verify that the MODE
is set for SNGL.
If it is not, press
SINGL
ENTR.
Press EXIT as
needed to
return to
SAMPLE
mode.
Verify that the RANGE is
set for 50.0
If it is not, toggle each
numeric button until the
proper range is set, then
press ENTR.
Verify that the UNIT
is set for PPM
If it is not, press
PPM
ENTR.
06864B DCN6314