Teledyne API – Model T300/T300M CO Analyzer Data Acquisition System (DAS) and APICOM
179
To define the REPORT PERIOD, follow the instruction shown in Section 7.1.4 then
press:
The SAMPLE PERIOD and REPORT PERIOD intervals are synchronized to the
beginning and end of the appropriate interval of the instruments internal clock.
If SAMPLE PERIOD were set for one minute the first reading would occur at the
beginning of the next full minute according to the instrument’s internal clock.
If the REPORT PERIOD were set for of one hour, the first report activity would occur
at the beginning of the next full hour according to the instrument’s internal clock.
EXAMPLE:
Given the above settings, if the DAS were activated at 7:57:35 the first sample would
occur at 7:58 and the first report would be calculated at 8:00 consisting of data points for
7:58, 7:59 and 8:00.
During the next hour (from 8:01 to 9:00), the instrument will take a sample reading
every minute and include 60 sample readings.
06864B DCN6314