Teledyne t320 Carbon Monoxide Alarm User Manual


 
Teledyne API – Model T300/T300M CO Analyzer Troubleshooting and Service
285
12.5.8.3. ANALOG OUTPUTS: CURRENT LOOP
To verify that the analog outputs with the optional current mode output are working
properly, connect a 250 ohm resistor across the outputs and use a voltmeter to measure
the output as described in Section 5.9.3.6 and then perform an analog output step test as
described in Section 12.5.8.2.
For each step the output should be wit
hin 1% of the nominal value listed in the table
below.
Table 12-11: Analog Output Test Function - Nominal Values Voltage Outputs
OUTPUT RANGE
2 -20 4 -20
NOMINAL OUTPUT VALUES
STEP % CURRENT V(250 OHMS) CURRENT V(250 OHMS)
1 0 2 mA 0.5V 4 1
2 20 5.6 1.4 7.2 1.8
3 40 9.2 2.3 10.4 2.6
4 60 12.8 3.2 13.6 3.4
5 80 16.4 4.1 16.8 4.2
6 100 20 5 20 5
06864B DCN6314