Teledyne t320 Carbon Monoxide Alarm User Manual


 
Teledyne API – Model T300/T300M CO Analyzer Calibration Procedures
217
EXIT returns
to the SETUP
Menu.
SETUP X.X DURATION:15.0 MINUTES
<SET SET> EDIT EXIT
SETUP X.X DURATION 15.0MINUTES
3 0 .0 ENTR EXIT
SETUP X.X DURATION:30.0 MINUTES
<SET SET> EDIT EXIT
Toggle buttons to set
duration for each iteration
of the sequence:
Set in Decimal minutes
from 0.1 – 60.0.
SETUP X.X CALIBRATE: OFF
<SET SET> EDIT EXIT
SETUP X.X CALIBRATE: OFF
ON ENTR EXIT
SETUP X.X CALIBRATE: ON
<SET SET> EDIT EXIT
Toggle button between
Off and ON.
SETUP X.X SEQ 2) ZERO
SPAN, 2:00:30
PREV NEXT MODE SET EXIT
Display show:
SEQ 2) ZERO–SPAN, 2:00:30
Sequence Delta Time
MODE Delta Days
CONTINUED FROM PREVIOUS PAGE
DELTA TIME
9.4.1.1. AUTOCAL WITH AUTO OR DUAL REPORTING RANGES MODES SELECTED
If the T300/T300M Analyzer is set for either the Dual or Auto reporting range modes,
the following three steps will appear at the beginning of the AutoCal setup routine:
SETUP X.X RANGE TO CAL: HIGH
<SET EDIT EXIT
EXIT returns to the
PRIMARY SETUP
Menu.
SETUP X.X SEQ 2) ZERO
SPAN, 2:00:30
PREV NEXT MODE SET EXIT
SETUP X.X RANGE TO CAL: LOW
LOW HIGH ENTR SETUP
SETUP X.X RANGE TO CAL: LOW
<SET EDIT EXIT
Note
In order to automatically calibrate both the HIGH and LOW ranges, you
must set up a separate sequence for each.
06864B DCN6314