Troubleshooting and Service Teledyne API – Model T300/T300M CO Analyzer
290
12.6.2. REMOVING/REPLACING THE GFC WHEEL
When removing or replacing the GFC Wheel it is important to perform the disassembly
in the following order to avoid damaging the components:
1. Turn off the analyzer.
2. Remove the top cover.
3. Open the instrument’s hinged front panel.
4. Locate the GFC Wheel/motor assembly. See Figure 3-6.
5. Unpl
ug the following electronic components:
The GFC Wheel housing temperature sensor
GFC Wheel heater
GFC Wheel motor power supply
SAFETY SHIELD
SYNCHRONOUS MOTOR
SOURCE ASSEMBLY
THERMISTOR
HEATER
Figure 12-15: Opening the GFC Wheel Housing
6. Remove the two (2) screws holding the opto-pickup printed circuit assembly to the
GFC Wheel housing.