Teledyne t320 Carbon Monoxide Alarm User Manual


 
Teledyne API – Model T300/T300M CO Analyzer Calibration Procedures
215
9.4.1. SETUP ACAL: PROGRAMMING AND AUTO CAL SEQUENCE
Note
If at any time an illegal entry is selected, (for example: Delta Days > 366)
the ENTR label will disappear from the control button.
To program the example sequence shown in Table 9-4, press:
CONTINUE NEXT PAGE
With STARTING TIME
SAMPLE RANGE = 50.0 PPM CO=XX.XX
< TST TST > CAL CALZ CZLS SETUP
SETUP X.X
CFG ACAL DAS RNGE PASS CLK MORE EXIT
SETUP X.X SEQ 1) DISABLED
NEXT MODE EXIT
SETUP X.X SEQ 2) DISABLED
PREV NEXT MODE EXIT
SETUP X.X MODE: DISABLED
NEXT ENTR EXIT
SETUP X.X MODE: ZERO
PREV NEXT ENTR EXIT
SETUP X.X SEQ 2) ZERO
SPAN, 1:00:00
PREV NEXT MODE SET EXIT
SETUP X.X TIMER ENABLE: ON
SET> EDIT EXIT
SETUP X.X STARTING DATE: 01
JAN
07
<SET SET> EDIT EXIT
SETUP X.X STARTING DATE: 01–JAN–02
0 4 SEP 0 8 ENTR EXIT
Toggle buttons to set
Day, Month & Year:
Format : DD-MON-YY
SETUP X.X MODE: ZERO
SPAN
PREV NEXT ENTR EXIT
06864B DCN6314