Teledyne T360 Carbon Monoxide Alarm User Manual


 
Operating Instructions Teledyne API Model T360/T360M Operation Manual
72
To view and clear warning messages:
SAMPLE SYSTEM RESET CO2 = 0.00
TEST CAL MSG CLR SETUP
Press CLR to clear the current
message.
If more than one warning is active, the
next message will take its place
Once the last warning has been
cleared, the analyzer returns to
SAMPLE mode
SAMPLE RANGE=500.000 PPM CO2 = 0.00
< TST TST > CAL MSG CLR SETUP
SAMPLE SYSTEM RESET CO2 = 0.00
TEST CAL MSG CLR SETUP
TEST deactivates warning
messages
MSG displays warning messages.
NOTE:
If the warning message persists
after several attempts to clear it,
the message may indicate a
real problem and not an artifact
of the warm-up period
Figure 4-3: Viewing and Clearing T360 WARNING Messages
4.3. CALIBRATION MODE
Pressing the CAL button switches the T360 into multi-point calibration mode. In
this mode, the user can calibrate the instrument or check the instruments
calibration with the use of calibrated zero or span gases.
If the instrument includes either the zero/span valve option or IZS option, the
display will also include CALZ and CALS buttons. Pressing either of these
buttons also puts the instrument into multipoint calibration mode.
CALZ is used to initiate a calibration of the zero point.
CALS is used to calibrate the span point of the analyzer. It is recommended
that this span calibration is performed at 90% of full scale of the analyzer’s
currently selected reporting range.
Because of their critical importance and complexity, calibration operations are
described in detail in Section 5 of this manual. For more information concerning
the valve options, see Section 3.6.
07272B DCN6552