Teledyne T360 Carbon Monoxide Alarm User Manual


 
Operating Instructions Teledyne API Model T360/T360M Operation Manual
144
4.13.4.5. AIN CALIBRATION
This is the sub-menu to conduct the analog input calibration. This calibration
should only be necessary after major repair such as a replacement of CPU,
motherboard or power supplies. Activate the ANALOG I/O
CONFIGURATION MENU (see Section 4.13.1), then press:
Exit at any time to
return to the main
DIAG menu
DIAG
A
NALOG I / O CONFIGURATION
PREV
NEXT
ENTR
EXIT
Continue pressing SET? until …
DIAG AIO
A
IN CALIBRATED: NO
< SET SET> CAL EXIT
DIAG AIO AIN CALIBRATED:YES
< SET SET> CAL EXIT
Instrument
calibrates
automatically
Exit to return to the
ANALOG I/O
CONFIGURATION
MENU
STARTING FROM ANALOG I / O CONFIGURATION MENU
DIAG AIO CALIBRATING A/D ZERO
CALIBRATING A/D SPAN
4.13.4.6. CONFIGURING ANALOG INPUTS (OPTION) CHANNELS
To configure the analyzer’s external analog inputs option, define for each
channel:
gain (number of units represented by 1 volt)
offset (volts)
engineering units to be represented in volts (each press of the touchscreen
button scrolls the list of alphanumeric characters from A-Z and 0-9)
whether to display the channel in the Test functions
07272B DCN6552