Teledyne T360 Carbon Monoxide Alarm User Manual


 
Teledyne API - T360/T360M, 360E/360EM Appendix A Menu Trees (05233C DCN6552)
A-22
Setup Variable Numeric
Units
Default
Value
Value Range Description
REF_SDEV_LIMIT mV 50 0.1–500
Reference detector standard
deviation must be below this limit
to switch out of startup mode.
3000 (not
used)
REF_SOURCE_LIMIT mV
Warnings:
1100–4800,
25–4800
3, 4, 15
1–5000
Reference source warning limits.
Set point is not used.
FACTORY_OPT BitFlag 512,
768
5
0–0x7fffffff
Factory option flags. Add values
to combine flags.
1 = enable dilution factor
2 = zero/span valves installed
4 = enable conc. alarms
8 = enable linearity adjustment
factor
16 = display units in
concentration field
32 = enable software-controlled
maintenance mode
64
3, 5
= span valve installed
128 = enable switch-controlled
maintenance mode
256 = compute only offset during
zero calibration
512 = 220 V A/C power
1024 = non-zero offset
calibration (linearity adjustment
must also be enabled)
2048 = enable Internet option
7
4096 = use “old” style numeric
data entry menus when editing
conc. table
8192 = locate high range and
zero cal. status outputs on relays
16384 = enable external analog
inputs
25
07272B DCN6552