Teledyne T360 Carbon Monoxide Alarm User Manual


 
Getting Started Teledyne API Model T360/T360M Operation Manual
60
3. Perform the Zero/Span Calibration Procedure
M-P CAL STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > ENTR CONC EXIT
M-P CAL STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > CAL SETUP
M-P CAL STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > ZERO CONC EXIT
EXIT returns to the main
SAMPLE display
A
CTION:
Allow zero gas to enter the sample port at the
rear of the instrument.
M-P CAL STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > SPAN CONC EXIT
M-P CAL RANGE = 500.000 PPM CO2 =XXX.X
< TST TST > ENTR SPAN CONC EXIT
M-P CAL RANGE = 500.000 PPM CO2 =XXX.X
< TST TST > ENTR CONC EXIT
Wait until STABIL
falls below 1.0 ppm.
This may take several
minutes.
The SPAN button now
appears during the
transition from zero to
span..
If either the ZERO or
SPAN buttons fail to
appear see Section 11
for troubleshooting tips.
SAMPLE* RANGE = 500.000 PPM CO2 =XXX.X
< TST TST > CAL SETUP
SAMPLE STABIL=XXX.X PPM CO2 =XXX.X
< TST TST > CAL SETUP
Set the Display to show the
STABIL test function.
This function calculates the
stability of the CO
x
measurement
Press ENTR to changes the
OFFSET & SLOPE values for the
CO
2
measurements.
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
A
CTION:
Allow span gas to enter the sample port at the
rear of the instrument.
The value of
STABIL may jump
significantly.
Wait until it falls back
below 1.0 ppm
This may take several
minutes.
Press ENTR to change the
OFFSET & SLOPE values for the
CO
2
measurements.
Press EXIT to leave the calibration
unchanged and return to the
previous menu.
07272B DCN6552