Teledyne API - T360/T360M, 360E/360EM Appendix A Menu Trees (05233C DCN6552)
A-29
ST_ZERO_CAL
10 + 13
6 1 = in zero calibration
0 = not in zero
ST_HIGH_RANGE2
16
7
1 = high auto-range in use (mirrors ST_HIGH_RANGE
status output)
0 = low auto-range
A status outputs, U24, J1017, pins 1–8 = bits 0–7, default I/O address 323 hex
ST_SYSTEM_OK 0 0 = system OK
1 = any alarm condition
ST_CONC_VALID 1 0 = conc. valid
1 = hold off or other conditions
ST_HIGH_RANGE 2 0 = high auto-range in use
1 = low auto-range
ST_ZERO_CAL 3 0 = in zero calibration
1 = not in zero
ST_SPAN_CAL 4 0 = in span calibration
1 = not in span
ST_DIAG_MODE 5
0 = in diagnostic mode
1 = not in diagnostic mode
ST_AUTO_REF
3
6
0 = in auto-reference mode
1 = not in auto-reference mode
7 Spare
B status outputs, U27, J1018, pins 1–8 = bits 0–7, default I/O address 324 hex
ST_AUTO_REF
2
0
0 = in auto-reference mode
1 = not in auto-reference mode
1–5 Spare
ST_CO2_CAL
7
6 0 = in CO
2
calibration
1 = not in CO
2
calibration
ST_O2_CAL
5
7 0 = in O
2
calibration
1 = not in O
2
calibration
Front panel I
2
C keyboard/touchscreen, default I
2
C address 4E hex
MAINT_MODE 5 (input) 0 = maintenance mode
1 = normal mode
LANG2_SELECT 6 (input) 0 = select second language
1 = select first language (English)
SAMPLE_LED 8 (output) 0 = sample LED on
1 = off
CAL_LED 9 (output) 0 = cal. LED on
1 = off
FAULT_LED 10 (output) 0 = fault LED on
1 = off
AUDIBLE_BEEPER 14 (output) 0 = beeper on (for diagnostic testing only)
1 = off
Relay board digital output (PCF8574), default I
2
C address 44 hex
RELAY_WATCHDOG 0
Alternate between 0 and 1 at least every 5 seconds to keep
relay board active
WHEEL_HTR 1 0 = wheel heater on
1 = off
07272B DCN6552