4.15.2.3. COMMAND SYNTAX
Commands are not case-sensitive and all arguments within one command (i.e. ID
numbers, keywords, data values, etc.) must be separated with a space character.
All Commands follow the syntax:
X [ID] COMMAND <CR>
Where
X is the command type (one letter) that defines the type of command.
Allowed designators are listed in Table 4-26 and Appendix A-6.
[ID] is the analy
zer identification number (Section 4.11.1). Example: the
Comm
and “? 200” follo
wed by a carriage return would print the list of
available commands for the revision of software currently installed in
the instrument assigned ID Number 200.
COMMAND is the command designator: This string is the name of the command
being issued (LIST, ABORT, NAME, EXIT, etc.). Some commands
may have additional arguments that define how the command is to
be executed. Press ? <CR> or refer to Appendix A-6 for a list of
available command designators.
<CR> is a carriage return. All commands must be terminated by a carriage
return (usually achieved by pressing the ENTER button on a
computer).
Table 4-26: Command Types
COMMAND COMMAND TYPE
C
Calibration
D
Diagnostic
L
Logon
T
Test measurement
V
Variable
W
Warning
4.15.2.4. DATA TYPES
Data types consist of integers, hexadecimal integers, floating-point numbers,
Boolean expressions and text strings.
Integer data are used to indicate integral quantities such as a number of
records, a filter length, etc. They consist of an optional plus or minus sign,
followed by one or more digits. For example, +1, -12, 123 are all valid
integers.
Hexadecimal integer data are used for the same purposes as integers. They
consist of the two characters “0x,” followed by one or more hexadecimal
digits (0-9, A-F, a-f), which is the ‘C’ programming language convention. No
plus or minus sign is permitted. For example, 0x1, 0x12, 0x1234abcd are all
valid hexadecimal integers.
Floating point numbers are used to specify continuously variable values such
as temperature set points, time intervals, warning limits, voltages, etc. They
07272B DCN6552