Teledyne T360 Carbon Monoxide Alarm User Manual


 
Teledyne API - T360/T360M, 360E/360EM Appendix A Menu Trees (05233C DCN6552)
A-25
Name
1
Message Text Description
WREARBOARD REAR BOARD NOT DET
Rear board was not detected during
power up.
WRELAYBOARD RELAY BOARD WARN
Firmware is unable to communicate with
the relay board.
WFRONTPANEL FRONT PANEL WARN
Firmware is unable to communicate with
the front panel.
WANALOGCAL ANALOG CAL WARNING
The A/D or at least one D/A channel has
not been calibrated.
Test Measurements
RANGE RANGE=50.0 PPM
3
CO RANGE=50.0 PPM
3, 7
D/A range in single or auto-range modes.
RANGE1 RANGE1=50.0 PPM
3
CO RANGE1=50.0 PPM
3, 7
D/A #1 range in dual range mode.
RANGE2 RANGE2=50.0 PPM
3
CO RANGE2=50.0 PPM
3, 7
D/A #2 range in dual range mode.
CO2RANGE CO2 RANGE=20 %
7
CO
2
range.
O2RANGE O2 RANGE=100 %
10
O
2
range.
STABILITY STABIL=0.0 PPM
3
CO STB=0.0 PPM
3, 7, 10
CO2 STB=0.0 %
7
O2 STB=0.0 %
10
Concentration stability (standard deviation
based on setting of STABIL_FREQ and
STABIL_SAMPLES).
RESPONSE
2
RSP=0.20(0.00) SEC
Instrument response. Length of each
signal processing loop. Time in
parenthesis is standard deviation.
COMEAS CO MEAS=4125.0 MV Detector measure reading.
COREF CO REF=3750.0 MV Detector reference reading.
MRRATIO MR RATIO=1.100 Measure/reference ratio.
AUTOZERO
4, 5
AZERO RATIO=1.234
Measure/reference ratio during auto-
reference.
SAMPPRESS PRES=29.9 IN-HG-A Sample pressure.
PURGEPRESS
9
PURGE=7.5 PSIG Purge pressure
VACUUM
8
VAC=6.8 IN-HG-A Vacuum pressure.
SAMPFLOW
6
SAMP FL=751 CC/M Sample flow rate.
SAMPTEMP
13
SAMPLE TEMP=26.8 C Sample temperature.
BENCHTEMP BENCH TEMP=48.1 C Bench temperature.
WHEELTEMP WHEEL TEMP=68.1 C Wheel temperature.
O2CELLTEMP
10
O2 CELL TEMP=50.2 C O
2
sensor cell temperature.
BOXTEMP BOX TEMP=26.8 C Internal box temperature.
BOXTEMP2
4
BOX TEMP2=29.6 C Internal box temperature #2.
OVENTEMP
11
OVEN TEMP=30.1 C Oven temperature
PHOTOTEMP PHT DRIVE=2500.0 MV Photometer temperature.
07272B DCN6552