Teledyne T360 Carbon Monoxide Alarm User Manual


 
Teledyne API Model T360/T360M Operation Manual Operating Instructions
137
4.13.4.1. ANALOG OUTPUT SIGNAL TYPE AND RANGE SPAN SELECTION
To select an output signal type (DC Voltage or current) and level for one output
channel, activate the ANALOG I/O CONFIGURATION MENU (see Section
4.13.4) then press:
DIAG AIO CONC_OUT_2 RANGE: 5V
SET> EDIT EXIT
DIAG ANALOG I / O CONFIGURATION
PREV NEXT ENTR EXIT
DIAG AIO AOUTS CALIBRATED: NO
< SET SET> CAL EXIT
FROM ANALOG I/O CONFIGURATION MENU
DIAG AIO OUTPUT RANGE: 5V
0.1V 1V 5V 10V CURR ENTR EXIT
To set the signal
level and type of
the selected
channel
DIAG AIO CONC_OUT_2:5V, CAL
< SET SET> EDIT EXIT
Press SET> to select the
analog output channel to be
configured. Press EDIT to
continue
Pressing ENT
R
records the new setting
and returns to the previous menu.
Pressing EXIT ignores the new setting
and returns to the previous menu.
DIAG AIO OUTPUT RANGE: 10V
0.1V 1V 5V 10V CURR ENTR EXIT
07272B DCN6552