Teledyne T360 Carbon Monoxide Alarm User Manual


 
Appendix C
Warranty/Repair Questionnaire
T360, M360E
(05235C DCN5798)
TELEDYNE API CUSTOMER SERVICE
Email: api-customerservice@teledyne.com
PHONE: (858) 657-9800 TOLL FREE: (800) 324-5190 FAX: (858) 657-9816
C-1
CUSTOMER: ____________________________________ PHONE: ______________________________________
CONTACT NAME: ________________________________ FAX NO: ______________________________________
SITE ADDRESS: __________________________________________________________________________________
SERIAL NO.: ____________________________________ FIRMWARE REVISION: __________________________
1. Are there any failure messages? ____________________________________________________________________
________________________________________________________________________________________________
________________________________________________________________________________________________
________________________________________________________________________________________________
________________________________________________________________________________________________
Please complete the following table:
PARAMETER DISPLAYED AS OBSERVED VALUE UNITS NOMINAL RANGE
Range
RANGE
PPM 0 -10, 0 - 2000 Ppm
Stability
STABIL
PPM 0.15 Ppm With Zero Air
CO
2
Measure
CO2 MEAS
mV 3600 – 4800 Mv
CO
2
Reference
CO2 REF
mV 1400 – 2000 Mv
Measure/Reference Ratio
MR RATIO
2.5 ± 0.02 W/ Zero Air
Pressure
PRES
INHG -2”Ambient Absolute
Sample Flow
SAMP FL
CC 800 ± 10%
Sample Temp
SAMPLE TEMP
ºC 48 ± 4
Bench Temp
BENCH TEMP
ºC 48 ± 2
Wheel Temp
WHEEL TEMP
ºC 68 ± 2
Box Temp
BOX TEMP
ºC Ambient + 7 ± 10
Photo Drive
PHT DRIVE
mV 250 Mv TO 4750 Mv
Slope
SLOPE
. 1.0 ± .3
Offset
OFFSET
0 ± 0.3
O
2
Sensor Temperature
3
O2 CELL TEMP
ºC 50 ± 5
Slope of O
2
Measurement
3
O2 SLOPE
Offset of O
2
Measurement
3
O2 OFFSET
Dark Cal Reference signal
REF DARK OFFSET
mV 125 ± 50 Mv.
Dark Cal Measurement Signal
MEAS DARK OFFSET
mV 125 ± 50 Mv
Electric Test
PPM 40 ± 2 Ppm
Values are in the Signal I/O
REF_4096_MV
4096mv±2mv And Must Be
Stable
REF_GND
0± 0.5 And Must Be Stable
2. Have you performed a leak check and flow check? ______________________________________________________
3. What is the sample flow & sample pressure with the sample in-let on rear of machine capped?
SAMPLE FLOW _________________________ CC SAMPLE PRESSURE ______________________ IN-HG-A
07272B DCN6552