147
4.13.6. Dark Calibration Test
The dark calibration test interrupts the signal path between the IR photo-detector
and the remainder of the sync/demod board circuitry. This allows the instrument
to compensate for any voltage levels inherent in the sync/demod circuitry that
might effect the calculation of CO
2
concentration. Performing this calibration
returns two offset voltages, One for CO
2
MEAS and on for CO
2
REF that are
automatically added to the CPU’s calculation routine. The two offset voltages
from the last calibration procedure may be reviewed by the user via the front
panel display.
To activate the dark calibration procedure or review the results of a previous
calibration, press:
SETUP X.X ENTER DIAG PASS: 818
8 1 8 ENTR EXIT
SAMPLE RANGE = 500.000 PPM CO2 =X.XXX
< TST TST > CAL SETUP
SETUP X.X PRIMARY SETUP MENU
CFG DAS RNGE PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG ALRM EXIT
Calibration runs automatically
Electric offset for Measurement signal
Electric offset for Reference signal
DIAG SIGNAL I / O
NEXT ENTR EXIT
Repeat Pressing NEXT until . . .
DIAG DARK REF DARK OFFSET: 0.0 mV
EXIT
DIAG DARK MEAS DARK OFFSET: 0.0 mV
EXIT
DIAG DARK DARK CAL 1% COMPLETE
EXIT
DIAG DARK DARK CALIBRATION ABORTED
EXIT
Display
tracks %
complete
DIAG DARK CALIBRATION
PREV NEXT ENTR EXIT
Exit returns
to the
previous menu
DIAG DARK CO2 DARK CALIBRATION
VIEW CAL EXIT