Teledyne T360 Carbon Monoxide Alarm User Manual


 
Troubleshooting and Service Teledyne API Model T360/T360M Operation Manual
254
8.5.8.5. CONTROL INPUTS – REMOTE ZERO, SPAN
The control input bits can be tested by the following procedure:
1. Connect a jumper from the +5 pin on the Status connector to the x5V on the
Control In connector.
2. Connect a second jumper from the ‘-‘ pin on the Status connector to the A pin
on the Control In connector. The instrument should switch from
SAMPLE
mode to ZERO CAL R mode.
3. Connect a second jumper from the ‘-‘ pin on the Status connector to the B pin
on the Control In connector. The instrument should switch from
SAMPLE
mode to
SPAN CAL R mode.
4. In each case, the T360 should return to SAMPLE mode when the jumper is
removed.
8.5.9. CPU
There are two major types of failures associated with the CPU board: complete
failure and a failure associated with the Disk on Module (DOM). If either of
these failures occur, contact the factory.
For complete failures, assuming that the power supplies are operating properly
and the wiring is intact, the CPU is faulty if on power-on, the watchdog LED on
the motherboard is not flashing.
In some rare circumstances, this failure may be caused by a bad IC on the
motherboard, specifically U57, the large, 44-pin device on the lower right hand
side of the board. If this is true, removing U57 from its socket will allow the
instrument to start up but the measurements will be invalid.
If the analyzer stops during initialization (the front panel display shows a fault or
warning message), it is likely that the DOM, the firmware or the configuration
and data files have been corrupted.
07272B DCN6552