5.9.3.4. AUTOMATIC GROUP CALIBRATION OF THE ANALOG OUTPUTS
IMPORTANT
IMPACT ON READINGS OR DATA
Manual calibration should be used for any analog output set for a
0.1V output range or in cases where the outputs must be closely
matched to the characteristics of the recording device.
Before performing this procedure, ensure that the AUTO CAL for
each analog output is enabled. (See Section 5.9.3.3).
To calibrate the outputs as a group with the AOUTS
CALIBRATION command, select the ANALOG I/O CONFIGURATION
submenu (see Section 5.9.3) then press: