Teledyne T801 Carbon Monoxide Alarm User Manual


 
Teledyne API T801 NDIR CO2 Analyzer Operation Manual Troubleshooting and Service
187
11.2.2.2. RELAY PCA STATUS LED S
D5 (Yellow) – CO2 Sensor Heater
D1 (RED)
Watchdog Indicator
Figure 11-4: Relay PCA Status LEDS Used for Troubleshooting
Table 11-4: Relay PCA Status LED Failure Indications
LED Color Function
Status When LED Lit
(Energized State)
Status When LED Unlit
(Default State)
D1
Red Watchdog Circuit
Cycles ON/OFF every 3 Seconds
under direct control of the analyzer’s CPU.
D2-D4 SPARE
D5
Yellow CO
2
Sensor Cell heater Heating Not Heating
D6
1
Green
D7
1
Green
D8
21
Green
D9
1
Green
D10
1
Green
D11 - 16 SPARE
1
Not Used
11.3. GAS FLOW PROBLEMS
When troubleshooting flow problems, it is a good idea to first confirm that the
actual flow and not the analyzer’s flow sensor and software are in error, or the
flow meter is in error. Use an independent flow meter to perform a flow check as
described in Section 10.3.4. If this test shows the flow to be correct, check the
pressure sens
ors as described in Section 11.6.8.