3.3.1.6. CONNECTING THE CONTROL INPUTS
If you wish to use the analyzer to remotely activate the zero and span calibration
modes, several digital control inputs are provided through a 10-pin connector
labeled CONTROL IN on the analyzer’s rear panel.
There are two methods for energizing the control inputs. The internal +5V
available from the pin labeled “+” is the most convenient method (Figure 3-10,
left). However, if full isolation is required, an external 5 VDC power supply
should
be used (Figure 3-10, right).
CONTROL IN
A B C D E F U
+
SPAN/ZERO CAL
CAL/SMPL MODE
CONTROL IN
A B C D E F U
+
-
+
5 VDC
Power Supply
Local Power Connections
External Power Connections
RANGE SELECTION
CO2 CAL
SPAN/ZERO CAL
CAL/SMPL MODE
RANGE SELECTION
CO2 CAL
Figure 3-10: Control Input Connector
Table 3-7: Control Input Signals
Input # Status Definition
Open/Closed Condition Description
A
CALIBRATION MODE or
SAMPLE MODE
Closed: initiates Calibration mode. Front panel display Mode field will read
CAL CO2 R or CAL CO2 ZR.
Open: initiates Sample (Measure) mode. Front panel display Mode field will
read SAMPLE.
B
REMOTE SPAN or REMOTE
ZERO CALIBRATION
Closed: initiates remote SPAN calibration mode as part of performing a low
span calibration. Front panel display Mode field will read CAL CO2 SR.
Open: initiates remote ZERO calibration mode. Front panel display Mode
field will read ZERO MODE.
C
RANGE2 or RANGE1
CALIBRATION
Closed: selects High Range for calibration (C2H).
Open: selects Low Range (C2L), default range in single range mode.
D
CO
2
SENSOR
CALIBRATION
Closed: Initiates CO
2
sensor calibration. Front panel display Mode field will
read CAL CO2 R or CAL CO2 ZR.
Open: Exits CO
2
sensor calibration and returns to SAMPLE mode.
E & F
SPARE
Digital Ground
The ground level from the analyzer’s internal DC power supplies (same as
chassis ground)
U
External Power input
Input pin for +5 VDC required to activate Pins A – F.
+
5 VDC output
Internally generated 5V DC power. To activate inputs A – F, place a jumper
between this pin and the “U” pin. The maximum amperage through this port
is 300 mA (combined with the analog output supply, if used).