Teledyne API T801 NDIR CO2 Analyzer Operation Manual Data Acquisition System (DAS) & APICOM
141
To define the REPORT PERIOD, follow the instruction shown in Section 7.1.4
then press:
The SAMPLE PERIOD and REPORT PERIOD intervals are synchronized to
the beginning and end of the appropriate interval of the instrument’s internal
clock.
If SAMPLE Period were set for one minute, the first reading would occur at
the beginning of the next full minute according to the instrument’s internal
clock.
If the Report period were set for one hour, the first report activity would occur
at the beginning of the next full hour according to the instrument’s internal
clock.
EXAMPLE:
Given the above settings, if DAS were activated at 7:57:35 the first sample would
occur at 7:58 and the first report would be calculated at 8:00 consisting of data
points for 7:58, 7:59 and 8:00. During the next hour (from 8:01 to 9:00), the
instrument will take a sample reading every minute and include 60 sample
readings.
07274B DCN6418