Teledyne API T801 NDIR CO2 Analyzer Operation Manual Getting Started
49
3.3.2.6. EXHAUST OUTLET
Attach an exhaust line to the analyzer’s EXHAUST outlet fitting. The exhaust
line should be:
PTEF tubing; minimum O.D ¼”
A maximum of 10 meters long.
Vented outside the T801 analyzer’s enclosure, preferably outside the shelter
or at least into a well-ventilated area.
Figure 3-17: T801 Internal Gas Flow
IMPORTANT
Leak Check:
Run a leak check once the appropriate pneumatic connections
have been made; check all pneumatic fittings for leaks using the
procedures defined in Section 10.3.3.
07274B DCN6418