To set individual ranges press the following menu sequence:
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
5.4.4. SETUP RNGE DIL: USING THE OPTIONAL DILUTION RATIO
FEATURE
This feature is an optional software utility that is used to compensate for any
dilution of the sample gas that may occur before it enters the sample inlet.
Typically this occurs in continuous emission monitoring (CEM) applications
where the sampling method used to remove the gas from the stack, dilutes the
sample.
To set up and use the dilution ratio option:
1. In the DIAG menu, use the 929 password and navigate to Factory Options
and enable the Dilution Ratio feature. (Press ENTER to save setting, and
then return to SETUP menu).