Teledyne API T801 NDIR CO2 Analyzer Operation Manual Setup Menu
65
5.4.2. ANALOG OUTPUT RANGES FOR CO
2
CONCENTRATION
The analyzer has several active analog output signals accessible through a
connector on the rear panel (see Figure 3-4).
Not Used
A
NALOG OUT
A1 A2
A
3
A
4
+ - + - + - + -
CO
2
concentration outputs
HIGH range when DUAL
or AUTO mode is selected
Test Channel
LOW range when DUAL
or AUTO mode is selected
Figure 5-1: Analog Output Connector Pin Out
All three outputs can be configured either at the factory or by the user for full
scale outputs of 0.1 VDC, 1VDC, 5VDC or 10VDC.
Additionally A1 and A2 may be equipped with optional 0-20 mA current loop
drivers and configured for any current output within that range (e.g. 0-20, 2-20,
4-20, etc.). (Please contact Sales for the Current Loop option) The user may also
adjust the signal level and scaling of the actual output voltage or current to match
the input requirements of the recorder or datalogger (See Section 5.9.3.9).
When the instrument is in
its default configuration, channels A1 and A2 (Dual or
Auto range) output a signal proportional to the CO
2
concentration of the sample
gas. (See Section 5.4.3). A3 is not used.
The output, l
abeled A4 is special. It can be set by the user (See Section 5.9.3.10)
to output several of the test functions accessible through the <TST TST> men
u
buttons.
07274B DCN6418