Teledyne API T801 NDIR CO2 Analyzer Operation Manual Setup Menu
67
5.4.3.1. RNGE MODE SNGL: CONFIGURING THE T801 ANALYZER FOR SINGLE
RANGE MODE
Single Range Mode (SNGL) is the default reporting range mode for the analyzer.
When the single range mode is selected (SNGL), all concentration outputs (A1,
A2) are slaved together and set to the same reporting range limits (e.g. 0- 20.00
%). The span limit of this reporting range can be set to any value within the
physical range of the analyzer.
Although the outputs share the same concentration reporting range, the electronic
signal ranges of the analog outputs may still be configured for different values
(e.g. 0-5 VDC, 0-10 VDC, etc; see Section 5.9.3.1)
To select SNGL range m
ode and to set the upper limit of the range, press:
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).